Turns out that you can, in fact, Telnet to instruments like Keysight DMM 34461A. For that specific meter, useful documentation is in manual section (paralleling the meter's menus):
There you can also enable/disable VXI-11, Sockets, and Web interface.
(The SCPI section of the manual rather unhelpfully does not point you to this info.)
Key nugget of info:
"The instrument Telnet port is 5024. Open SCPI sessions on the Telnet connection by entering: telnet IP address 5024"
So the very simplest "Hello World" exercise is as follows:
- Connect the meter ethernet port to the LAN.
- Check the Utility Menu > I/O Configuration > LAN Settings page for the IP address it obtained via DHCP, or to set IP address.
- Use a terminal program (for example PuTTY) set for Telnet, with port 5024.
- Start a Telnet session to the IP address of the meter.
The meter responds with a prompt. At this point you can enter SCPI commands. For example *IDN? returns an identifying string.
Comments
This Telnet access is, of course, not secure. But then neither are the web services nor VXI-11 (that cooperates with VISA). All of these can be turned off if you want.
There are undoubtedly merits to using VISA, and possibly also an IVI instrument driver (that runs on top of VISA), which further civilizes and speeds up access to particular instruments. But this Telnet path gives at least bare bones access without having to deal with any complexities of vendor-provided (or NI-provide) piles of software and corresponding libraries that you may or may not need, and might have conflicts with each other.